Postby tonizg » Wednesday 9. November 2016, 20:11

Dear Sirs,

My question:
Q1: SPECTRAN NF 5035 "Logger function": In "Logger" function/menu: Is it possible to use other numbers for "FILEId" and what are those numbers (in manual is stated that only possible number is "1000")?

from manual "SPECTRAN-NF_EN.pdf" (Rev 1.33 EN 14.03.2013):
"
7.21 Logger (Invoke recording/data logger)
...
After activating, the device will query the following parameters:
“Count” = Enter the number of logs (recorded events)
“Time” = Enter the duration of the pauses between the individual logs, in seconds.
“FILEId” = Enter the program number used to store the recorded data.
This number is later used to download the recorded data using our PC?software. ATTENTION! Currently, our PC?software only supports program number “1000“!
"

Best regards,
tonizg
tonizg
 
Posts: 9
Joined: Sunday 2. October 2016, 23:39

Re: SPECTRAN NF 5035 "Logger function": "FILEId" more than 1

Postby mm_dev » Thursday 10. November 2016, 15:51

You can use any number above 1000, however the PC software to actually see the data only recognizes file ID 1000 as logger data by default.
mm_dev
 
Posts: 284
Joined: Wednesday 4. May 2011, 07:43