HF & NF SPECTRAN: which actions will not erase calibration?

Aaronia Spectrum Analyzer (SPECTRAN NF, SPECTRAN HF, SPECTRAN XFR, SPECTRAN X, SEPCTRAN 19", Spectran V5)

HF & NF SPECTRAN: which actions will not erase calibration?

Postby tonizg » Monday 17. October 2016, 16:35

Dear Sirs,
I have been informed that SPECTRAN devices (NF & HF) were calibrated in factory (for precise measurements levels) and also from manual and forum posts that is usually needed to send them to AAronia at least every 2 years (or on MCS will "popup" message that device should be calibrated if user wants to be sure that measurements are accurate).
During those 2 years, I am interested in answer to question:

Q1: Does following actions on SPECTRAN affect on calibration of device (user doesn't need to send SPECTRAN back to AAronia for calibration)? :
a) if user executes “Factor” under the “Setup” menu
b) if battery inside SPECTRUM completely discharge itself (because device wasn't used for several months)
c) runing some MCS functions [similar to a) but from MCS menu:Spectran->Factory reset
d) if user do FIRMWARE update by himself (following procedure from manual)

Thanks.
Regards,
tonizg
tonizg
 
Posts: 9
Joined: Sunday 2. October 2016, 23:39

Re: HF & NF SPECTRAN: which actions will not erase calibrati

Postby Moderator » Friday 21. October 2016, 15:11

a) if user executes “Factor” under the “Setup” menu

Keeps the calibration
b) if battery inside SPECTRUM completely discharge itself (because device wasn't used for several months)

Keeps the calibration
c) runing some MCS functions [similar to a) but from MCS menu:Spectran->Factory reset

Keeps the calibration
d) if user do FIRMWARE update by himself (following procedure from manual)

Keeps Calibration if not noted otherwise.
Moderator
 
Posts: 1036
Joined: Tuesday 20. April 2010, 13:24
Location: Aaronia AG


Return to Spectrum Analyzer