How to correct the reading error?

Hilfe und Informationen zur Firmware der SPECTRAN NF- Serie
Firmware related help and information for the SPECTRAN NF- series

How to correct the reading error?

Postby Leandro » Thursday 28. January 2016, 22:33

Hi, I’ve checked the magnetic field reading of the meter by placing it in a known 50Hz magnetic field from 1µT to 1mT and I detected that the measuring error went from -1dB to -1.8dB. I checked each axis separately and measured the same. The configuration was the following:

• Center freq: 50Hz
• Span: 30Hz.
• RBW: 3Hz.
• Attenuation: 20dB till 100µT and then 30dB.

I wanted to know whether there is a way to correct the reading by software or other method. Thanks.
Leandro.
Leandro
 
Posts: 2
Joined: Thursday 21. January 2016, 19:05

Re: How to correct the reading error?

Postby Moderator » Saturday 30. January 2016, 10:04

If your unit shows wrong values send it to our lab for calibration (would cost 199,905 Euros plus shipping).
You can adjust the readings in the MCS software in the calibration menue (see manual).
Did you rise the sample time to maximum (3000mS)? Otherwise you will get slightly to low readings what seems to be the case.
Moderator
 
Posts: 1036
Joined: Tuesday 20. April 2010, 13:24
Location: Aaronia AG

Re: How to correct the reading error?

Postby Leandro » Friday 12. February 2016, 13:36

Thank you, I changed the Sweep Time to 3000ms, set a calibration off-set of 0.7dB and I verified the readings for 50Hz, different B field intensities and the reading was very accurate.
Leandro
 
Posts: 2
Joined: Thursday 21. January 2016, 19:05

Re: How to correct the reading error?

Postby Moderator » Friday 12. February 2016, 21:47

Thanks for the feedback, just as expected :D